    Oh,and of course I would never violate my precious new baby by sticking any kind of probe inside!
    Even in the quest for knowledge. I just wish the 2215 interface could give me the temp as well as the current and output .

  2. #12
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,031

    Default

    [QUOTE=HD58PHD;141817]Oh,and of course I would never violate my precious new baby by sticking any kind of probe inside!


    cue one, " crude alien probing joke", go cue...

    If it looks like this:

    http://secure2.data-comm.com/servlet...-Blower/Detail

    or This (a smallish version of Rotron Patriot)

    http://candhsurplus.com/product/dcfans/comair48-.jpg

    You might be on the right course...

    BTW, if you can stop the fan with your fingers not requiring a trip to the ER, you probably have the wrong fan

    Steve

    Steve
    Qui habet Christos, habet Vitam!
    I should have rented the space under my name for advertising.
    When I still could have...

  3. #13
    Join Date
    Feb 2010
    Location
    Utah
    Posts
    96

    Default

    I should be good to go then,the fan on it is rated 170 cfm.

    Click image for larger version. 

Name:	P4070052..JPG 
Views:	8 
Size:	676.4 KB 
ID:	15885
    Thanks again
    Mike

  4. #14
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,031

    Default

    [QUOTE=HD58PHD;141932]I should be good to go then,the fan on it is rated 170 cfm.


    EBM/PABST makes good stuff

    Your fine in the fan department... How long is the duct length? length is bad, but each bend is a 30+% reduction in airflow...

    Steve
    Qui habet Christos, habet Vitam!
    I should have rented the space under my name for advertising.
    When I still could have...

  5. #15
    Join Date
    Feb 2010
    Location
    Utah
    Posts
    96

    Default

    O.K. now don't laugh,I know it's kinda ghetto,but it works and it's compact!Click image for larger version. 

Name:	P1060034..JPG 
Views:	27 
Size:	691.3 KB 
ID:	15888

  6. #16
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,031

    Default

    Quote Originally Posted by HD58PHD View Post
    O.K. now don't laugh,I know it's kinda ghetto,but it works and it's compact!Click image for larger version. 

Name:	P1060034..JPG 
Views:	27 
Size:	691.3 KB 
ID:	15888

    Ghetto, my foot... Well Done! Proper sized fan, short air path, and proper tapered nozzle. Add a short section of rubber for vibration damping and you'd be all set.

    Steve
    Qui habet Christos, habet Vitam!
    I should have rented the space under my name for advertising.
    When I still could have...