Results 1 to 5 of 5

Thread: Setting up CT scanners

  1. #1
    Join Date
    Nov 2010
    Location
    Denmark
    Posts
    244

    Default Setting up CT scanners

    I recently got a set of CT6800HPs on ebay (without drivers).

    Now I also have a pair of CT MicroMax 670 scanner drivers on the way, so need to get them up and running with the scanners.

    Already got the 670 tuning guide from the nice people at CT - have been using it as bedtime reading.

    Now the question - The tuning guide is pretty generic, and not written with laser projection in mind.

    Tuning guide says Open loop tuning -> notch filter tuning -> coarse tuning -> fine tuning

    Do I switch to the ILDA pattern after the coarse tuning, or is the fine tuning really needed first?

    Another thing - The drivers support Class 0 and Class 1 tuning - which one is most suitable for our purposes?

    Thanks,
    Thomas

  2. #2
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,066

    Default

    Fine tuning/ILDA is a tossup
    You do need the Notch if the notch matches the 6800s. If not, the notch is a PITA, and can usually be bypassed.
    Class 0 is preferred for speed, but either way will work.
    Class 0 will need more work to get right.

    Can I have a copy of the microamp manual? I know you can't post it.
    Steve

  3. #3
    Join Date
    Nov 2010
    Location
    Denmark
    Posts
    244

    Default

    Quote Originally Posted by mixedgas View Post
    Fine tuning/ILDA is a tossup
    Exactly what I was expecting

    Quote Originally Posted by mixedgas View Post
    You do need the Notch if the notch matches the 6800s. If not, the notch is a PITA, and can usually be bypassed.
    Okay, haven't received the drivers yet, so I don't know which notch filters they have.
    Will find the resonance frequency of the scanners, and use the notch if it is in range... If not, I'll just run without.

    Quote Originally Posted by mixedgas View Post
    Class 0 is preferred for speed, but either way will work.
    Class 0 will need more work to get right.
    What kind if speed penalty will I get from class 1? The extra precision is tempting for graphics

    Thomas

  4. #4
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,066

    Default

    I'm guessing ~10% less speed.
    Did your manual cover multiple parallel notches?
    Did the manual cover finding resonances? If not, I have one that does.



    Steve

  5. #5
    Join Date
    Nov 2010
    Location
    Denmark
    Posts
    244

    Default

    Quote Originally Posted by mixedgas View Post
    I'm guessing ~10% less speed.
    Did your manual cover multiple parallel notches?
    Did the manual cover finding resonances? If not, I have one that does.
    Oh okay, class 0 it is then

    Manual covers both single and dual notch filters - resonance testing as well

    You have a pm

    Thanks, Thomas
    Last edited by Badpip; 07-09-2012 at 17:30.

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •