Thread: Coherent Star C : H/E low reservoir

  1. #1
    Join Date
    Dec 2008
    Posts
    35

    Default Coherent Star C : H/E low reservoir

    Hello,

    A friend has a Coherent Innova Star C (mixed gas). It was working very well until this week. Now, the remote displays "Heat exchanger low reservoir". In the manual, it is said that it's because the reservoir of the optionnal heat excahenger (Water/water exchanger or chiller) is too low.

    BUT this laser doesn't have a heat exchanger, it's connected directly to the sanitary cold water, it's open loop. So it shouldn't display this message...

    There is a connector named J17 to connect an optionnal heat exchanger to the power supply. This connector is used to power the heat exchanger AND to control the water level in the exchanger. Does someone know the pinout and voltage of this connector ? I could try to put a shunt or the right voltage to inform the power supply that the water level is OK ?

    Thank you for your help

  2. #2
    Join Date
    Sep 2012
    Posts
    64

    Default

    It has to have a flow sensor somewhere. When using open loop the water tends to foul up the sensor. If your lucky you can pull it and clean it off. White vinigar in a 50/50 with water works good to get the gunk off.

  3. #3
    Join Date
    Dec 2008
    Posts
    35

    Default

    I don't think it because of the water flow rate. There is an other fault if the flow rate is too low (fault messsage "low water flow").

  4. #4
    Join Date
    May 2009
    Location
    Sacramento, California
    Posts
    609

    Default

    The person who would proberly know is DSLi-Jon. Hes a Purelight Guru.
    Will there be three phase!!!!

  5. #5
    mixedgas's Avatar
    mixedgas is online now Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,012

    Default

    Quote Originally Posted by Vidal Wolf View Post
    The person who would proberly know is DSLi-Jon. Hes a Purelight Guru.
    Call/email scientifc laser service @ the Coherent factory. Ask them which two pins the flow jumper goes on, on the external connector.
    They will not talk to you unless you give them a head serial number and a psu serial number. They usually will send you a manual in PDF form.

    Every Coherent heat exchanger that I've worked on has a contact closure for "cooling valid", and I suspect you need to jump those two pins on the laser.
    The data will be in the pdf manual.

    Steve
    Last edited by mixedgas; 01-04-2013 at 07:51.

  6. #6
    Join Date
    Sep 2010
    Location
    Munich Germany
    Posts
    112

    Default

    Hi,
    if you open the powersupply you can find the right pinout for the water sensor, you only need to use a wire to connect two pins together
    There is also a setting. If your power supply has a build in service eprom, you can put on dip switch number 7 and you can change the setting via remote...

  7. #7
    Join Date
    Dec 2008
    Posts
    35

    Default

    Somebody told me that I have to put only the switch n°7 ON to enter in the service mode. I should also put ON only the switch n°7 INSIDE the power supply. Is it right ?

  8. #8
    Join Date
    Dec 2008
    Posts
    35

    Default

    The method to go into the service menu with dip switch n°7 worked. The guy changed the settings of the heat exchanger from "Yes" to "No".
    But now, there is an other default :"bad autofill cal". Maybe all settings are erased because of a default from the capacitor to store the settings ?

  9. #9
    mixedgas's Avatar
    mixedgas is online now Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,012

    Default

    Oh, you have 300 series type power supply technology. Not what I thought you had. I've had that on UV midsize argons at work.

    IS the power supply the origional one for the head he has?

    Bad Autofill Cal is/was bad news on that series. It means the head eeprom no longer matches the tube voltage curve when the PSU runs its self test. You can run the Cal if you have FULL access to service mode ( You may/may not need a service chip) and the right remote, provided your on a very clean source of line power (Not a phase converter). If it won't calibrate after running the service program, the tube pressure needs adjusted by a laser reprocessing company.

    It can take 45 minutes to three hours or more of running to calibrate on a 300. I'll have to check my copy of the service manual if I still have it. The 300 procedure will be similar to Purelight but I would be contacting Coherent. You have one generation newer then what I have worked on. I learned this one the hard way after rewindowing and repumping a UV tube myself. I had to run the tube pressure up and down on the pumping station quite a bit before I found where the PSU was happy.

    Perhaps Matratzki knows, as I have not worked on your more recent generation. My knowledge is not up to date on this one.

    Since I do not have physical acess to the laser, I cannot help you on this one.

    Good Luck,



    Steve
    Last edited by mixedgas; 01-06-2013 at 14:33.

  10. #10
    Join Date
    Dec 2008
    Posts
    35

    Default

    I think that the tube is OK. It worked 2 weeks ago. I think that the problem is from the capacitor that power the EEPROM to store the settings. If the capacitor is broken, all settings are erased and that can explain why we have to change the heat exchanger setting from "Yes" to "No". There wasn't a heat exchanger before...

    Is there a manual to install and configure this laser for the first time ? First commissioning with service settings.