    Default LEXEL 295 POWER SUPPLY AC CONTACTOR Problem of AC config problem? Please help

    I have a lexel 295 supply running a 95 head. I really need a service manual to know if the AC 3 phase power is properly configured for the current coming in. At this time when I push the power momentary button the AC contactor hums and does not lock on. Is it bad? or could this be indicative of a power configuration problem? Any information would be helpful. Thank You!

  2. #2
    mixedgas's Avatar
    mixedgas is online now Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,012

    Default

    That is a indication of a specialized interlock fault unique to 295s. I'm away from my schematics library till late monday evening. You'll need a tap chart and some other data. Please dont try to run it till you have the expanded charts.

    Steve

  3. #3
    Join Date
    Feb 2013
    Location
    Tampa, Florida
    Posts
    7

    Default

    Thank you Steve for responding quickly! Might this be Steve R?

    Remember the lexel sent to you packed so you could drop it
    Off a building and it would survive? (2006)

    I am just getting back to the drawing board so to speak.

    I will post 1 to 2, 2 to 3, 1 to 3, and ground to 1,2and 3 voltage in a couple of hours

    I had a feeling that the power supply was that smart.

    Thank you!

    Jon (in Tampa)

    [/QUOTE]

  4. #4
    mixedgas's Avatar
    mixedgas is online now Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,012

    Default

    OK, the schematic attached is a 95. Follow the interlock chain, find the differences and eliminate them. This is the easiest way I have found to deal with that external interlock issue.. You may have a second big relay in your PSU with a second filter cap, its a start booster. If you have that, we may need to adapt things a bit.

    You do have a remote interlock jumper plug installed, correct? If not, that will stop it cold.

    I'll scan in a tap chart tomorrow.

    When I get home I'll highlight the chain using Windows Paint. We have Linux machines at work and do not really have a good image editor.

    Steve R.
    Attached Thumbnails Attached Thumbnails LEXEL95SCHEM.jpg  


  5. #5
    Join Date
    Aug 2008
    Location
    the hills
    Posts
    983

    Default

    I'll look, there is a possibility i have a tap chart on my external, i'll post in a sec, if I can find it