Page 1 of 2 12 LastLast
Results 1 to 10 of 15

Thread: Scored a Laserscope YAG!

  1. #1
    Join Date
    Apr 2008
    Location
    DC/VA metro area
    Posts
    425

    Default Scored a Laserscope YAG!

    This is what everyone loves to hear :-)

    Where exactly does the beam shoot out from? KIDDING!

    Picked up a laserscope YAG recently. It's a 800 series, built in heat exchanger.

    Did some research before hand and have continued to do more research on them.

    I'm a good ways from trying to fire it up.

    Been reading the documentation on ?DJ Allens? site. He has what appears to be the class manual from a service tech class on the units. Very very very informative. Same with the video on youtube.

    Has anyone attempted to build a microcontroller based computer replacement for these? I saw the video on youtube where someone was using the project box style switch boxes, and a meter. Seems like a good fit for a controller. I'd be willing to collab on this and share the results.

    The earlier project on here where someone split the PSU / cooling / head apart seems ripe. Holy HELL is the thing heavy. I guess I should of expected it, but yea. We laid it down in a pick up and to get it out we had to tie rope to the wheels -- then to a pole and drive away to slide it out of the truck :-) I was eyeballing it for places to shed weight.

    I've been researching DI water. DI water is required, correct? Not distilled? I'm going to poke around locally and see if there is a source for DI water.

    I assume that a series of peltier coolers wouldn't have enough umph to chill a water loop instead of the radiator. 5000W is ~17,000 btu/hr. Hmmm peltiers are only 100-300 btu it seems. I assume the lamp runs solid and the only pulse is a result of the q-switch gating the output. 17,000 btu seems way more than a radiator could dissipate so my number is way off there I assume.

    I looked to see if there are any newer more modern power supplies that are smaller for the lamp, but nothing jumped out.

    Also going to research safety eyewear that matches the system.

    Neat machine though!

    Are conversions still in use much in the light show world or is everyone diode?

  2. #2
    Join Date
    Nov 2007
    Location
    SoCal / San Salvador / NY
    Posts
    4,018

    Lightbulb

    Heya tel -

    Some 'unsolicited food for thought'..

    Quote Originally Posted by telmnstr View Post
    ..I'm a good ways from trying to fire it up.
    Yes, please. You need pulsed-laser system safety fundamentals session, first and foremost. This is *not* a 'learn by trial and error' kind-of laser system. Even if you didn't 'mind being blind', the 'typical' first-temptation is 'show the new hot rod off to-friends', etc, and as the saying goes, 'friends don't let friends blind other friends'.

    So yea - we needa tawk... I'd happily offer a 'discussion-session', but I am booked-solid till after July 6th (..I'm only 'here', right now, waiting for 3D-renderings / video to crunch, etc..) Check with 'P'fessor Steve' (mixedgas) He (may) have more time avail, than I, meantime...

    Quote Originally Posted by telmnstr View Post
    Been reading the documentation on...
    ..Mk, yeah... (insert facepalm) Don't believe everything you can 'glean off the interwebz'.. Much of that information is *incorrect*, or, at the very-least, not 'optimal'.. Sure, the 'OEM Service manuals' are "helpful" and good to have, great, fine. But, other info is just flat-out wrong. I can't tell you how many 'help calls' I've gotten from people, saying 'well, I followed the info on that site, and my stuff aren't-working.. helllp!..' ... You may think '...yeah, but you could 'learn a lot from a dummy'.. but, this is not one of those cases.. So, yeah... training, *first*, then, help towards 'proper conversion path'... Which will also-be based on your 'intentions' for-use (ie: public-shows, of any-sort, v/o a Var, etc.. = Ø help.

    Quote Originally Posted by telmnstr View Post
    Has anyone attempted to build a microcontroller based computer replacement for these?...
    Yes, the 'Godfather of Green', Laserman532, *pioneered* it, other Co's have built very-respectable versions (ie: 'StellaRay'..) and there are even PL-members that have done-so... Speaking for-myself / Co. tho, I have always felt more 'comfy' flinging-around that kind of power, with 'dumb'-old hard-wired keyswitches and toggles...

    Quote Originally Posted by telmnstr View Post
    I was eyeballing it for places to shed weight.
    There's a (useless for anything but surgery-stable input-power..) 150# 1:1 transformer in the base, but.. like I said.. 'baby-steps'..

    Quote Originally Posted by telmnstr View Post
    DI water is required, correct? Not distilled?
    Nope, not-needed. But the 'correct' answer (..and 'why' it is correct..) will be given after 'class'..

    Quote Originally Posted by telmnstr View Post
    5000W is ...
    ...4.5kW is min. and the *entire* cooling-loop must be maintained as-designed.. As-to 'why', again, is related to the above-answer, yours for the 'price of admission'..

    Quote Originally Posted by telmnstr View Post
    Also going to research safety eyewear that matches the system.
    Also, something better-covered 'in class'.. Eyeware - by-itself - can lead to a 'false-sense of security', and actually 'create' other hazards... Not to say they are 'not needed', but.. there's a 'seasoned-approach' to working with these systems, that just 'buying a pair of 1064/532 glasses' does-not teach...

    Quote Originally Posted by telmnstr View Post
    Are conversions still in use much in the light show world or is everyone diode?
    Sometimes, there's just nothing-else that will-do... And, in the not-too-distant future, you are (imho..) going to see a 'resurgence of use' of these, when a certain 'augmenting product' hits the market... Some of the biggest Sho co's still maintain an arsenal of them, because in certain-circumstances / applications, there's just (still..) no 'topping' them.. And there's a -specific-reason-, beyond just 'sheer-power' / color...

    Anyhoo, please take this all in the 'spirit in-which it is intended' - to-help But, please also do-know that safely, reliably, and responsibly-operating these systems is *not* just a case of 'all you need is a dollar and a key'... Much, much more to making a success of these than 'the score'...

    Talk to you more, asap..
    j
    Last edited by dsli_jon; 06-18-2013 at 22:42.
    ....and armed only with his trusty 21 Zorgawatt KTiOPO4...

  3. #3
    Join Date
    Apr 2008
    Location
    DC/VA metro area
    Posts
    425

    Default

    Very cool. I'm aware that when driven in q-switch mode the output technically looks like something of a saw wave, and power measurements will only show an average of that.

    As far as replacement controller, I'd still have a hardware keyswitch to turn the whole thing off, then a second keyswitch for the q-switch driver if that can be done.

    The only manuals I've really looked at were the OEM service manuals and some of the schematics for the bypass controller type stuff.
    I read a few old threads that I could find with information (such as early 700s have unreliable/failure prone power supplies.)

    My first use is basically a static shot into the sky during an event I help run, and I've already talked to the FAA trying to figure out if we had any special cases here since we have two airports that are commercial, then a few that are .mil. They said all they need is the form. Of course my side needs variance as well. And I recently found out from another group in the area they were denied by the FAA (but they were sitting under a flight path I think.)

    Interesting regarding the 1:1 iso transformer. Normally that is done as a requirement of the power supply and for safety? Most switchers I think have them on board so external ones aren't needed. If it's there as some sort of medical equipment requirement to protect the hospital electrical system from noise or something that's great if I can shed it.

    Last night I was looking around a bit and I see things like the Neslab CFM-33 chiller, which if it has a pump something like that could take the place of all the water handling in the 'scope, but with non-conductive water.

    I have owned a 3 phase argon in the past (6 watt SP 2011, that I actually sold on here to a fella out of PA.) I do realize this is a special dangerous beast!

  4. #4
    Join Date
    Feb 2011
    Location
    New Hampshire
    Posts
    3,513

    Default

    My first use is basically a static shot into the sky during an event I help run, and I've already talked to the FAA trying to figure out if
    This statement is VERY significant. You really need to spend some quality time discussing your plans with Jon before going any further. REALLY! The fact that this laser is 50x as powerful as the most powerful laser pointers and 7x as powerful as your argon is not as significant as the Q switching. Each of these short (300 nsec pulses) packs 10-20 K Watt and it explodes things like optics and retinal blood vessels.

  5. #5
    Join Date
    Apr 2007
    Location
    Costa Rica
    Posts
    523

    Default

    Quote Originally Posted by telmnstr View Post
    ...Interesting regarding the 1:1 iso transformer. Normally that is done as a requirement of the power supply and for safety? Most switchers I think have them on board so external ones aren't needed. If it's there as some sort of medical equipment requirement to protect the hospital electrical system from noise or something that's great if I can shed it...
    Careful! If your ALE lamp supply's cable shield is grounded to the box with the strain relief, be sure to isolate cable shield from the strain relief/chassis ground or you WILL blow up the ALE when you eliminate the isolation transformer and try to power it up.
    ale10.jpg

  6. #6
    Join Date
    Apr 2008
    Location
    DC/VA metro area
    Posts
    425

    Default

    Quote Originally Posted by Photonbeam View Post
    Careful! If your ALE lamp supply's cable shield is grounded to the box with the strain relief, be sure to isolate cable shield from the strain relief/chassis ground or you WILL blow up the ALE when you eliminate the isolation transformer and try to power it up.
    ale10.jpg
    Crazy! Thanks for the heads up on that! That is the kind of stuff that I was thinking would creep up as an issue when removing the 1:1. That seems kind of sketchy.

    My unit is actually in a different state from me at the moment (it's in a warehouse with a group I'm involved in up in the DC area) and I haven't brought it down to Southeastern Virginia yet so I haven't inspected it much past a few minutes :-/

    @planters - oh I'll definitely continue to do research on it and talk to Jon beforehand. I'm a ways away from powering it up.

  7. #7
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,114

    Default

    Believe me on this. The internal software is needed until you have mastered 'scope diagnostics. Do NOT modify the cooling system, keeping the rod at a constant temperature is very important for thermal lensing.

    FAA wants a variance. Even todays "meh!" CDRH is going to look long and hard at issuing a varience for a 'scope only user. Especially in the DC Aerospace area.

    These drill anodized aluminum. They etch uncoated metal, which blows back crud on the optics and kills the laser. It is a very high peak power system. Just finding a good beam dump can be problematic.

    Very little in Allen's on-line manuals is useful. The manual is set up to meet the bare minimum FDA requirements that require enough data for an end user to theoretically be able to maintain the unit. In reality medical manufacturers are not inclined nor required to put everything you need to know in the basic service manual. I spent a year working for a medical equipment manufacturer, and the regulatory system really requires you to attend a factory course for a reason. You follow the exactly the steps in the Lscope manual, you will probably end up at a dead end. You will find yourself making phone calls to find out why the system just shut down. That is done for safety reasons industry wide.

    Allen was/is showing poor judgement and mental illness in his attitude and safety practices with regard to these units. So rule him out for advice.

    Tuning it correctly requires two very good power meters. One sensitive one and one brute force one. Both are in use at the same time. Advanced users can get by with one really good LPM, but that comes with some risks.

    Beware of the modified so called "service" firmware that is floating around for these things for show use. It has a serious safety glitch in it. The one ON-LINE mods schematic is down right flawed. On some models it will kill the doubling crystal very quickly from lack of giant pulse supression.

    Going from "NO PEAK POWER" CW lasers to one of these is awakening in how little one knows. As a service engineer, I teach pulsed laser users on a daily basis in the nanosecond and and picosecond regimes. Just knowing when you can remove the safety glasses requires serious effort, thinking, and measurements. The precautions required need a great deal of discipline. I'm good at what I do, however, My employer does not like some of my lasershow habits, with regards to beam viewing, and he has a point. Just because I do it intrensically and instinctively does not mean some one who is watching me will do it correctly down the road.

    You need room, you need power, you need beam dumps and goggles and fire extinguishers, , you need tooling, You need spare lamps, crystals, and spare mirrors, and you need training on the system. Anything else is an invitation to disaster, or a lot of expensive repair parts. In the field, you need a trained assistant or two, you should not work on these alone. In the field, when working at these power levels, I either kick everyone out of the room, or EVERY move is discussed and briefed beforehand.

    Please don't point me to all these youtube videos where unskilled people have not followed that route. Those systems are coasting along based on their inherent design stability. Which degrades even when setting still, inoperative.

    Your already dreaming of modifying this system when you have no idea how it works, and that is a BAD thing.

    I'm not going to teach this over the phone. It is mentally intensive and dangerous. You need to hire Jon.

    The Per'fesser's services for pulsed laser show lasers are 1500$ weekend plus airfare, plus you have to attend one of the three or four one day laser safety courses I approve of, before hand. Ie Greg Makhov, Casey Stack, Rockwell, or James in England. Plus waver of liability. There is also an evaluation of you as a reponsible person. For CW I come out for 1/3rd of that,(or often just to make friends) so that should tell you this is not a light undertaking.

    I'll see if I can find a decent on line safety manual for you, but beware, this thing bites. If you do not have glasses on and you accidently view a terminated spot from even say 20 feet, the afterimages are massive, painful, debilitating, and can last for hours, if you do not do damage. NOT A TOY!



    Steve
    Last edited by mixedgas; 06-20-2013 at 07:10.
    Qui habet Christos, habet Vitam!
    I should have rented the space under my name for advertising.
    When I still could have...

  8. #8
    Join Date
    Aug 2008
    Location
    the hills
    Posts
    983

    Default

    Make certain to listen to these gentlemen (and Pat-Laserman532 too, if he chimes in)very carefully. They know what they are talking about, they have been there and done that. They will not BS you, but at the same time, you will not get spoon-fed either. Allen's site is not a good place for proper info.

    I would highly recommend proper LS safety glasses, but don't let the glasses/goggles make you complacent. A reflection from the inside surface of the glasses can blind you.

    And *NO* the open sky is not suitable beam termination (especially in the USA).

    I'll reiterate what they said: these are not toys.

  9. #9
    Join Date
    Nov 2007
    Location
    SoCal / San Salvador / NY
    Posts
    4,018

    Lightbulb

    ^^ Everything Steve said, with two exclamation points.

    Quote Originally Posted by telmnstr View Post
    ...They said all they need is the form.
    Heh - that 'reply' is an inch-wide and a mile deep.. First, it's not 'also need to get a Variance' - without a valid-Variance, in place, already, they won't even consider the 7140 app - you will be denied-permission, before getting to first-base. (...Not a really-great way to 'get your relationship started' with the FAA, btw.. )

    Second, speaking of 'relationship', yes - the FAA 'knows these systems', and knows them well.. They (also) may not approve this - simply because - you have no prior-history of successful-filings with them.. and asking to shoot a static-beam into their airspace with The Big Bad Green Monster is your first?? Mmmyeah, I'm not the FAA, but.. I'm gotta go with 'declined', sorry to opin.

    ..Anyhoo, not tryin to 'kick you in las pelotas', be some d-bag 'laser cop', but.. You need to know - unequivocally - if you want *any* 'help' with this whole process / path to-success, it's going to-be 'conditional'... And first-up - gotta resist subscribing to the whole 'I JUST GOT A DICK-UPGRADE!!!' attitude-thing that so-many guys seem to latch onto, when buying these off Fleabay, etc... Owning / converting / operating a former-surgical KTP system - *especially* - in any-sort of public-setting, requires a very-serious / astute attitude, methodical-approach to the training / homework necessary, and.. baby-steps. '..For my first-trick, I am going to shoot a beam into airspace' is not that 'best-practices' / baby-steps approach, dig?

    Hire the P'Fessor, or, we can talk later in the year.. I've got several-people in line ahead of ya, and they must come first... But, as with the others, your 'patience will be amply-rewarded' / pay-off, when everything works, works-well, and you don't melt eyeballs / down-planes, etc...

    DSLS_KTPCntrlr_v4.JPGDSLS_KTPCntrlr_v3.2_Lit.jpg, http://www.photonlexicon.com/forums/...161#post262161 blah, blah, blah...

    cheers!
    j
    ....and armed only with his trusty 21 Zorgawatt KTiOPO4...

  10. #10
    gashead's Avatar
    gashead is offline Admin Verified: Best Accent Ever(Tm)
    Join Date
    May 2009
    Location
    Burntisland, Scotland
    Posts
    899

    Default

    Quote Originally Posted by dsli_jon View Post
    YAG Tease.

    :P .
    "The horizon leans forward offering you space, to place New steps of change"

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •