Thread: Bio Rad / American Laser -> Need Help Identifying the Interlocks

  1. #11
    Join Date
    Jun 2008
    Location
    Orange County, CA
    Posts
    236

    Default

    I spent more time with the BioRad laser today and discovered that there is another interlock in the 8-pin connector at the back of the laser head. Two of the pins are connected to a dead man switch in the laser case. Once I placed a jumper between these two pins, the power supply turned on!!!

    As soon as the power supply turned-on, an alarm went off in the case. I thought something was connected wrong but after about 20 seconds, the laser emitted a blue-white light!!!

    It turns out, that the laser will emit without anything in the power supply "Interface" connector.

    Now with the laser on, I have more information:

    The 8-pin connector on the laser head has two pins for 28VDC to power the fan. I am currently running the laser with an external centrifugal fan that I use with my Uniphase argon laser.

    The laser head has a switch labeled "Low" and "High."

    I probed the "Interface" connector and measured the following voltages with the switch on "High" and "Low":

    Click image for larger version. 

Name:	Interface Voltages.jpg 
Views:	3 
Size:	17.8 KB 
ID:	40924

    Now that the laser works, can you help me with the interface connections to adjust the power level?

    Thank you for your help!
    Last edited by turbohead; 11-17-2013 at 08:06. Reason: Added Interface voltages with High and Low laser output