Eric,
One thing to do if your afraid and have the proper gear:
I start tuning with the amps on current limited bench power supplies, with the positive and negative rails set to ~200 mA limits at +/- 24 Volts. It gives me peace of mind when starting on unknown tunings or amps that have been adjusted by the incompetent. The positive rail will draw a current a bit higher then the negative rail, this current is the position sensor LED current plus about 10 mA. Position sensor LEDs, depending on the Galvo specs, are usually set between 10 and 60 mA.
Circuit Specialists sells nice 0-3 Amp 0-24V Chinese bench supplies with adjustable current limiting and LCD displays for about 99$ each. I buy them when they go on sale for about 59$ each. I would not use them "RAW" to test laser diodes, they have some surges as you set the current, but for testing amps and diode drivers they are the Cat's Meow.
Quiescent current on the scanner amps can be as much as 100 mA per supply rail on each axis, when not scanning. This will rise to around 250 mA for Cambridge products scanning at small angles and climb from that point. This depends on the model of Galvo. As you start to scan, the current does not rise much until your waveform gets complex or the angle goes wide. The PEAK currents can be 3-4-5 amps, which is why I suggested external capacitors be added to power supplies in projectors.
Steve