I brought a 95 card to work, I'll take pics of where to start on my break. Have a 85 and 95 in the lab now. Sorry for the delay.
AC Test point settings for Cathode Transformer:
Filament Transformer
Turn on "Line" circuit breaker. Measure AC voltage between TP5 and TP6 on front panel. SWITCH OFF "Line" circuit breaker before changing switch S10:
Voltage less than 116 VAC: Set S10 to position "A".
Voltage between 116 and 126 VAC: Set S10 to position "B".
Voltage more than 126 VAC: Set S10 to position "C".
If less than 108 or greater than 135, consult manufacturer.
Each tube has its own test sheet - use those settings instead if available.