Get one end really close using the HENE, then SLIGHTLY loosen (~0.010 to 0.020 Inch) the two OUTER screws on the circular mirror mount. Rock and scan the loosened mirror mount by hand against the end plate. Note where the laser flashes and adjust the mirror end plate as needed. This compensates for the index of refraction of the Brewster windows and the parallelism error of the mirrors.. Usually I start with the HR for this task.
The Omni optics have a slight wedge by design, this is what is mainly vexing you. This ensures the ghost beam is not collinear with the main beam.
The first time someone tries a short cavity aircooled, usually it takes them 2-4 hours to get it right.
Make sure the Brewsters are VERY clean.