    Hi,

    I have a Lexel 88 Ion laser head, which I am using here in GB. It came with a stepdown transformer,- which steps down all of 10 volts! We are obviously on 230v, the Exciter is rated 220v, what I want to know is how sensitive the exciter would be to have the extra 10v to its AC Input, as the step down transformer weighs a tonne and if I dont need it i would rather not bother with it. However I am afraid to try it without until I have an expert opinion....

    Regards,

    Alex

  2. #2
    Join Date
    Jun 2006
    Location
    Charlotte, NC
    Posts
    1,125

    Default

    The two lexel 88's I have, have can buck or boost between 202 and 245. The stepdown isnt needed so long as you set the legs right on TP3

  3. #3
    Join Date
    Jan 2008
    Posts
    49

    Default

    Hi thanks for your reply,

    I'm not too familiar with the Power suppies on the Lexel's, where abouts do I find that TP, is it one of the TAPs on the transformer inside (not much of it appears to be labeled you see

    Regards

  4. #4
    Join Date
    Jun 2006
    Location
    Charlotte, NC
    Posts
    1,125

    Default

    sorry I meant tb2, it is located under the roughly 2x6in service panel on the top of the PS, removable by 2 screws

    snips taken from here
    http://groups.google.com/group/alt.l...883f8abb138b44

    Tap Chart for Lexel-88 Power Supply
    (This is basically the same for the Lexel-85.)
    Buck/Boost Transformer


    Measured Voltage Orange Wire Blue Wire Action
    ------------------------------------------------------
    190-203 VAC TB2-1 TB2-3 Boost
    203-215 VAC TB2-2 TB2-3 Boost
    215-227 VAC TB2-1 TB2-1 Even
    227-238 VAC TB2-2 TB2-1 Buck
    238-245 VAC TB2-3 TB2-1 Buck


    Note: Orange wire labeled "O" and blue wire labeled "B" on redrawn
    schematic.

    Turn on "Line" circuit breaker. Measure AC voltage between TP5 and TP6
    on front panel. SWITCH OFF "Line" circuit breaker before changing
    switch S10:


    Voltage less than 116 VAC: Set S10 to position "A".
    Voltage between 116 and 126 VAC: Set S10 to position "B".
    Voltage more than 126 VAC: Set S10 to position "C".
    If less than 108 or greater than 135, consult manufacturer.
    Each tube has its own test sheet - use those settings instead if