Not a z-fold pro, but here's what I'd do, after my DIY expirience.
Assuming You haven't been moving/removing any optics recently - I'd:
-try fine tuning the SAM optic (dead end of the short arm). If You're watching the spot position on crystal (532nm/1,06 goggles!), this can be done with q-sw gated, other tuning better proceed at CW operation.
-if max power cannot be pushed up higher SAM way - remove the KTP crystal (remember the arrow position on it's side) and inspect crystal faces under low power microscope. I suppose the power limit comes from to some extent degraded AR coatings, so there's little You can do without replacement. One could try moving the spot on the crystal to a side or up/down (done using SAM and RAM optics).
If You have good 532nm/1,06 goggles You can notice brighter dots on crystal face during high power q-sw operation. These are the craters in crystal surface killing lasing in this spot area.
If You suspect that LAM optics may have been moved recently, I'd try to walk the laser with LAM / SAM pair. But remember to keep the spot on the crystal!
I woudn't recommend rough re-aligning the z-fold, as it's easy to loose lasing. If You've haven't been replacing any optics the chances it's misaligned seriously is low.
How's the lamp hours ?
Piotr.K