This happens quite a bit when you work with industrial equipment and office equipment. The solution is a proper ground bus system. You were actually pretty lucky. As mentioned you probably didn't only get 17VAC as it generally takes a bit more to overcome skin resistance to a point you feel it unless there were other factors. High current, high frequency equipment love inductive coupling to metal optical benches and the capacitance across the plastic computer case could have been significant. My guess is your computer was acting like a cap with a bleeder resistor across it. When you grabbed the detector head, the reactance was either lower than the head to the table (maybe you were a better ground) or the reactance was higher and you only get a portion of a big zap. Put a scope on the detector head and power up the laser. Even if you don't see a clean climb in voltage, if you are seeing a large amount of noise you have a ground problem.