Thread: Apple + Krypton Laser = ZAP!!

  1. #1
    Join Date
    Jan 2007
    Location
    Cupertino, California
    Posts
    2,130

    Laser Warning Apple + Krypton Laser = ZAP!!

    So yesterday when I was at work, we were aligning the argon and krypton lasers.

    I had my power meter hooked to my laptop and then the laptop was plugged into the wall with the standard power adapter. The laptop was a first generation unibody MacBook and the power meter was a Coherent LM-30 detector hooked to a Coherent SSIM module hooked with USB to the computer. The computer was on an insulated surface and the only thing touching the optical table the lasers were on was the detector head on a metal stand.

    When I reached down to adjust the position of the detector head, I got a nice shock. We measured the voltage between the table and the head when they wern't touching and we got about 17V AC.

    Keep in mind that both the laser system and the charger for the laptop were grounded properly. Anyone have any idea why this is happening? I was kinda worried about the power meter and computer because of the voltage, but there didn't seem to be any problems.

    Anyways, the moral of the story is that the Apple charger does not appear to be floating from the main AC line.

    Additionally, because this thread is useless without a picture, here's a picture of the Argon at full power going into a large Brewster cut ruby crystal.
    Attached Images Attached Images
    Last edited by Laser Ben; 06-28-2009 at 00:06.
    CLICKY!!!

    Admin: In the immortal words of Captain Planet: YOU HAVE THE POWER
    Admin: (To quit being a bitch)

  2. #2
    Join Date
    Nov 2007
    Location
    Cairns, Australia
    Posts
    1,896

    Default

    I doubt it was the 17V that got you, it may have been wayyy more before you touched it. Maybe your mains isnt properly grounded? Is your workbench grounded too? Possible that something in the laser could have been inducing a current to your bench, especially since it contains high voltage.

  3. #3
    Join Date
    Dec 2008
    Location
    Southport, UK
    Posts
    2,746

    Default

    Something was floating. I probably would have checked the table to a known good ground and the power head to a know good ground. My Fujitsu laptop psu floats at about 118v when its not gounded (on a 240v line). Point being; if both points were grounded to a low resistance ground, with low inductance cables ie not 30 feet of coiled up extension cable, then there shouldn't have been any significant PD between them.

    I'm not fully clued up on your wiring schemas over there, but if this happened on a ring main system; I'd be checking the integrity of that, straight after checking the integrity of all internal equipment connections and power cords.

    We have this thing (pain in the arse) called P.A.T. over here, were every 6 months, anything with a mains plug on the end has to be Portable Appliance Tested. Esentially an insulation flash, ground bond connection and ground leakage test. This period is now 3 months for construction work (a joke since all construction appliances have to be 110v CTE).
    http://img62.imageshack.us/img62/3985/laser.gif

    Doc's website

    The Health and Safety Act 1971

    Recklessly interfering with Darwin’s natural selection process, thereby extending the life cycle of dim-witted ignorami; thus perpetuating and magnifying the danger to us all, by enabling them to breed and walk amongst us, our children and loved ones.





  4. #4
    Join Date
    Jun 2009
    Location
    Nebraska
    Posts
    168

    Default

    Quote Originally Posted by Things View Post
    I doubt it was the 17V that got you, it may have been wayyy more before you touched it. Maybe your mains isnt properly grounded? Is your workbench grounded too? Possible that something in the laser could have been inducing a current to your bench, especially since it contains high voltage.

    I agree, there must be several thousand volts of potential there if you say you got a "good shock". And since it was a "good shock" I can't say that it was a typical scenario where you were walking around the la in your slippers by shuffling your feet. I think the best source for this high voltage must most likely come from your PSU. There is def something floating on that table, cause your table just acted like a single plate on a large capacitor! That discharge couldv'e been worse! Worse for your health!

    If you can't find the problem with your PSU, then a nice ground for your table should take care of it for now, UNLESS your PSUs output rail is what is somehow getting to your table. You don't want to directly short your PSU.

  5. #5
    Join Date
    Dec 2008
    Location
    Southport, UK
    Posts
    2,746

    Default

    Just a thought; was this an AC shock or a single discharge? If it was a single discharge you might have built up a net static charge from the HV field and discharged yourself through the bench.
    http://img62.imageshack.us/img62/3985/laser.gif

    Doc's website

    The Health and Safety Act 1971

    Recklessly interfering with Darwin’s natural selection process, thereby extending the life cycle of dim-witted ignorami; thus perpetuating and magnifying the danger to us all, by enabling them to breed and walk amongst us, our children and loved ones.





  6. #6
    Join Date
    Jun 2009
    Location
    Nebraska
    Posts
    168

    Default

    Quote Originally Posted by Doc View Post
    Just a thought; was this an AC shock or a single discharge? If it was a single discharge you might have built up a net static charge from the HV field and discharged yourself through the bench.
    Good thought, so... wait..

    Ben, you said you MEASURED 17V AC? It was a continuous AC?

    uhh.....

  7. #7
    Join Date
    Dec 2008
    Location
    Southport, UK
    Posts
    2,746

    Default

    The measured 17v might be totally irrelevant to the shock he experienced?
    http://img62.imageshack.us/img62/3985/laser.gif

    Doc's website

    The Health and Safety Act 1971

    Recklessly interfering with Darwin’s natural selection process, thereby extending the life cycle of dim-witted ignorami; thus perpetuating and magnifying the danger to us all, by enabling them to breed and walk amongst us, our children and loved ones.





  8. #8
    mixedgas's Avatar
    mixedgas is offline Creaky Old Award Winning Bastard Technologist
    Infinitus Excellentia Ion Laser Dominatus
    Join Date
    May 2007
    Location
    A lab with some dripping water on the floor.
    Posts
    10,119

    Default

    Check the green wire ground on the tandem pair, coherent uses a lot of aluminum in that frame and oxide builds up on AL the instant it hits the air, On the bottom of the association channel sled the cavities are mounted on is a big green wire that loosens up ever once in a while, make sure the lock washer is biting into the AL, and then ohm the ground all the way back to the power source.

    I really doubt its the coherent because it set up tp be operating room safe, ie hospital "green dot" wiring, which means unless you removed the intrinsic grounding features or your ina miswired building, its something else.

    Also check the grounds to the tube magnet cases, theya re brass and supposed to be green wire grounded. Other then that, having worked on more then a dozen dual head coherent sleds, I've never felt a tingle, and I've mounted them on optical tables that were earth bonded in ways you would not believe. Remember the sled was set up in such a way,that a patient/doctir/nurse on a grounded OR table in a properly wired OR could never even get a current of 30 microamps from the laser.

    Which leads me to my next point. It may not be the sled.

    which leaves, the ground wiring in your building, or the next culprit, which is the most likely one.

    The days of the totally transformer isolated wall wart are gone. Be aware that modern switching wall warts and laptop power supplies have a .1 uF "safety cap" from the cold "isolated" side to the line.

    There way a thread on the sci.electronics.design newsgroup a while back, and a bunch of serious EEs and ETs checked their switching wall warts and almost all, found they could get a 20V ac tingle to damp skin, current limited by the capacitive reactance combined with the bodies impedence.

    So ground the snot out of everything, study up on "star" or "home run" single point grounding.

    And may the schwartz be with you.

    Steve
    Qui habet Christos, habet Vitam!
    I should have rented the space under my name for advertising.
    When I still could have...

  9. #9
    Join Date
    Jul 2008
    Location
    Miami
    Posts
    506

    Default

    Oh ya !! ground potential difference sucks ! Ive got 70 vac shock from cold ground to hot ground before and it sucks !!
    " MANUFACTURER OF HIGH QUALITY MICRO LASER COMPONENTS" !!
    http://www.microlaserlabs.com/

  10. #10
    Join Date
    Nov 2007
    Location
    Michigan
    Posts
    97

    Default

    This happens quite a bit when you work with industrial equipment and office equipment. The solution is a proper ground bus system. You were actually pretty lucky. As mentioned you probably didn't only get 17VAC as it generally takes a bit more to overcome skin resistance to a point you feel it unless there were other factors. High current, high frequency equipment love inductive coupling to metal optical benches and the capacitance across the plastic computer case could have been significant. My guess is your computer was acting like a cap with a bleeder resistor across it. When you grabbed the detector head, the reactance was either lower than the head to the table (maybe you were a better ground) or the reactance was higher and you only get a portion of a big zap. Put a scope on the detector head and power up the laser. Even if you don't see a clean climb in voltage, if you are seeing a large amount of noise you have a ground problem.
    The Frothy Chimp
    Cynic Extraordinaire
    Back off man, I'm a scientist

    Good whiskey, fine cigars, long legged women and blues guitar.
    That's what I like.