STEVE PLEASE CONFIRM THIS METHOD.
I know nothing of the HGM-5 or what ever it is...BUT...
With all sp glass plasma tubes if a tube was high pressure, and we were positive that the magnet was working properly. We would "flash the cathode" (usually due to the customer over filling it) (it also sometimes works with water vapor contamination from long term storage) It was also a last ditch effort to avoid the customer having to buy a new PT. (note: will not work on a PT that is up to air)
it would work like this...
We would disconnect the cathode wires from the terminal block and connect the cathode wires to a set of jumper cables. We would remove the battery from our rental car and drag it into the lab.
You connect the negative cable to the battery using the jaws (+ or - dosnt really matter) to one side of the cathode. Then with the water flowing you touch the positive cable to the terminal post. The cathode will start to glow then glow brighter then all of a sudden you will see the cathode FLASH, FLASH, FLASH, FLASH!!! (like a strobe light) Then you remove the battery connection and allow to cool.
Each time we did this we would burn off about 10 volts of PT pressure. We found it was a good idea to do it twice then hook back up the laser and run it normally to measure how much gas you burned off.
it works well with SP tubes...disclaimer...attempt at your own risk
Pat B
laserman532 on ebay
Been there, done that, got the t-shirt & selling it in a garage sale.