MPE is 2.5mW/cm2 which equates to a measurement of 1mW through a 7mm diameter circular aperture - reasonably assumed to be the size of a human dilated pupil.
Note that the beam must fully fill the sensor.
Pi*r^2
3.1415 * (3.5mm^2) = 38.5mm^2
2.5mW * 0.385 = 0.96mW (or thereabouts)
My meter sensor is 8x8mm (64mm^2) so 1.6mW is my measured limit in the absence of any other control to limit maximum exposure, such as a scanfail.
250ms is assumed to be the maximum exposure time, based on the aversion response (blinking/looking away), but if you can control this by other means, and significantly reduce it, then you can use multiples of MPE.
But this is just skimming the surface really, you need to explore this i far more depth than I've got time to regurgitate right now. I agree, James Stewarts LVR safety course is really very good, although I must say that I had a really solid grounding in what was involved from reading this forum, so you can pick up a lot here.